Re: EEG and EOG help

Click for full size
Since nobody has responded so far ... please have a look at what this guy has done to get decent EOG output with cheap silver electrodes.

»makeprojects.com/Project/Eyeboar···/1#s8781

P.S. I recall you occasionally visit Ottawa. My wife is an ophthalmic medical technologist just in case you need a professional EOG and any other eye checkups done.


SparkChaser
Premium
join:2000-06-06
Downingtown, PA
kudos:3
Reviews:
·Verizon FiOS
Thanks lutful, yes that is the one that got me started. It's not for me but for someone that is paralyzed.

I can't get a decent signal out of his circuit. He has some strange values in his notch filter which doesn't make it work very well. I modded the filter part. Still not getting good signals. There are a few circuits floating around the net. Pretty much all the same functional blocks. I'm starting to think I need better electrodes and placement.
--
--
--
"Facts do not cease to exist because they are ignored." - Aldous Huxley

lutful
... of ideas
Premium
join:2005-06-16
Ottawa, ON
kudos:1
Click for full size
said by SparkChaser:

There are a few circuits floating around the net. Pretty much all the same functional blocks


I think some of those circuits came from ECGs. EOG signals are smaller in amplitude and lower in frequency.

You may recall there was an ECG discussion last year where I mentioned there are dedicated ICs which incorporate most of the front-end signal processing. Please have a look at Analog Devices.

said by SparkChaser:

I'm starting to think I need better electrodes and placement.

Most probably skin oil has accumulated on them. You could clean up with alcohol and apply some gel.

EOG-based computer goggles actually place the electrodes almost ideally. The reference electrode smack in the middle of the forehead and the left/right electrodes equal distance apart from that on the temples.

public

join:2002-01-19
Santa Clara, CA
said by lutful:

said by SparkChaser:

There are a few circuits floating around the net. Pretty much all the same functional blocks


I think some of those circuits came from ECGs. EOG signals are smaller in amplitude and lower in frequency.

This will not work. You will never match the resistors in the input for acceptable cmrr.
Also need low pass on the input to prevent cm voltage driving the amp into the rails.

lutful
... of ideas
Premium
join:2005-06-16
Ottawa, ON
kudos:1
Click for full size
downloadeog_project.pdf 3,084,368 bytes
said by public:

This will not work

Have a look at the captured EOG waveforms from a crude breadboard implementation of that "textbook" circuit in this project report from which I copied that diagram.

I agree that the input can be improved to filter out noise above 100Hz.

public

join:2002-01-19
Santa Clara, CA
said by lutful:

said by public:

This will not work

Have a look at the captured EOG waveforms from a crude breadboard implementation of that "textbook" circuit in this project report from which I copied that diagram.

I agree that the input can be improved to filter out noise above 100Hz.

Well working means working not just in a shielded room.
Bio signals are typically uV range, in common mode noise of several V. So max cmrr is essential.

lutful
... of ideas
Premium
join:2005-06-16
Ottawa, ON
kudos:1
Click for full size
said by public:

Well working means working not just in a shielded room.

Of course I agree with you ... but popular "eye-control" science projects using that basic design somehow work with buzzing lights, wifi hotspots and cell phones in high school gyms.

Anyway, in this thread and a previous ECG/EEG thread, I mentioned there are integrated circuit front-ends for ExG signal pickup from companies like Analog Devices. They take care of component matching and provide very good CMRR.

said by public:

Bio signals are typically uV range, in common mode noise of several V. So max cmrr is essential.

I commented on actual EOG amplitude/frequency earlier and mentioned that it is quite different than ECG. Here are two charts which provide characteristics of various bio signals and the main considerations for amplifier design.

You will see that EOG is actually quite robust requiring mainly good DC response, reasonably low drift and of course very high gain. CMRR and noise (even 50/60Hz) is less of a concern with EOG than other bio signals.

However, I agree that you could come up with a much better (discrete) front-end circuit for EOG than the two examples posted so far in this thread. My suggestion is to search for an IC that does the job.